Which invasive blood pressure monitoring transducer artifact occurs when the natural frequency of the monitoring system is close to the frequency of…
Practice question from Monitoring & CPR (NEET PG Preparation Anaesthesia), with the correct answer and a step-by-step solution.
hard · MCQ · Q-52F636
Which invasive blood pressure monitoring transducer artifact occurs when the natural frequency of the monitoring system is close to the frequency of the pulse wave, resulting in a falsely elevated systolic pressure reading?
- A.UnderdampingCorrect answer
- B.Overdamping
- C.Air bubble in tubing
- D.Excessive tubing length
Solution
Underdamping in an arterial line system causes resonance, leading to an exaggerated waveform with a falsely high systolic pressure and presence of extra (spurious) oscillations. Overdamping results in a sluggish waveform with a falsely low systolic pressure.
More from Monitoring & CPR
- A 45-year-old male is undergoing general anesthesia. During mechanical ventilation, the end-tidal carbon dioxide (EtCO2) suddenly drops fro…
- During CPR in an adult patient with an advanced airway in place, what is the recommended ventilation rate according to current AHA guidelin…
- Which of the following pulse oximetry phenomena is based on the principle that the pulsatile component (AC) of light absorption corresponds…
- A patient in the ICU suffers ventricular fibrillation cardiac arrest. After high-quality CPR and the first delivery of a biphasic defibrill…
Practise more Anaesthesia
Sign up free for adaptive practice on this chapter, with instant feedback and hints on every question.
Start practising freeFree to start · No credit card needed